Static Characteristics
AU
: May−06,11,13,14,15,16,18,19, Dec.−04,08,09,ll,12,14,15
As
mentioned earlier, the static characteristics are defined for the instruments
which measure the quantities which do not vary with time. The various static
characteristics are accuracy, precision, resolution, error, sensitivity,
threshold, reproducibility, zero drift, stability and linearity.
1. Accuracy
It
is the degree of closeness with which the instrument reading approaches the
true value of the quantity to be measured. It denotes the extent to which we
approach the actual value of the quantity. It indicates the ability of
instrument to indicate the true value of the quantity. The accuracy can be
expressed in the following ways : −
1) Accuracy as 'Percentage of Full Scale
Reading' : − In case of instruments having uniform scale,
the accuracy can be expressed as percentage of full scale reading. For example,
the accuracy of an instrument having full scale reading of 50 units may be expressed
as ± 0.1 % of full scale reading. From this accuracy indication, practically
accuracy is expressed in terms of limits of error. So for the accuracy limits
specified above, there will be ± 0.05 units error in any measurement. So for a
reading of 50 units, there will be error of ± 0.05 units i.e.± 0.1 % while for
a reading of 25 units, there will be error of ± 0.05 units in the reading i.e.
± 0.2 %. Thus as reading decreases, error in measurement is ± 0.05 units but
net percentage error is more. Hence, specification of accuracy in this manner
is highly misleading.
2)
Accuracy as 'Percentage of True Value':− This is the best
method of specifying the accuracy. It is to be specified in terms of the true
value of quantity being measured. For example, it can be specified as ± 0.1 %
of true value. This indicates that in such cases, as readings get smaller,
error also gets reduced. Hence accuracy of the instrument is better than the
instrument for which it is specified as percent of full scale reading.
3)
Accuracy as 'Percentage of Scale Span':− For an instrument, if
amax is the maximum point for which scale is calibrated, i.e. full scale
reading and amin is the lowest reading on scale. Then (amax − amin)
is called scale span or span of the instrument. Accuracy of the
instrument can be specified as percent of such scale span. Thus for an
instrument having range from 25 units to 225 units, it can be specified as ±
0.2 % of the span i.e. ± [(0.2/100) × (225 − 25)] which is ± 0.4 units error in
any measurement.
4)
Point Accuracy :− Such an accuracy is specified at
only one particular point of scale. It does not give any information about the
accuracy at any other point on the scale. The general accuracy of an instrument
cannot be specified, in this manner. But the general accuracy can be specified by
providing a table of the point accuracy values calculated at various points
throughout the entire range of the instrument.
Thus
the accuracy, in whatever way it may be specified, gives the equipment accuracy
with a particular set up and other conditions and does not include any personal
accuracy.
2. Precision
It
is the measure of consistency or repeatability of measurements.
Key Point It denotes the
closeness with which individual measurements are departed or distributed about
the average of number of measured values.
Let
us see the basic difference between accuracy and precision. Consider an instrument
on which, readings upto 1/1000 th of unit can be measured. But the instrument has
large zero adjustment error. Now every time reading is taken, it can be taken
down upto 1/1000 of unit. So as the readings agree with each other, we say that
the instrument is highly precise. But, though the readings are precise upto 1/1000
th of unit, the readings are inaccurate due to large zero adjustment error.
Every reading will be inaccurate, due to such error. Thus a precise instrument
may not be accurate.
Thus
the precision means sharply or clearly defined and the readings agree among
themselves. But there is no guarantee that readings are accurate.
An
instrument having zero error, if calibrated properly, can give accurate readings
but in that case still, the readings can
be obtained down upto 1/1000 th of unit only. Thus accuracy can be improved by
calibration but not the precision of the instrument.
Consider
another example. A known weight of 100 grams is measured by an instrument. Five
times, the weight has been recorded. The readings obtained are 103, 104, 105,
103, 105. The average indicated value is 104 grams. Hence the maximum deviation
from the average reading is ±1 gram in 100 grams actual value. Thus, the scale
of the instrument can be calibrated to read ±1 gram. But what about the
accuracy ? The readings are not accurate. The accuracy of the instrument is
only (105 − 100/100)% i.e. 5 % . Thus there is a precision of ± 1 % but the
accuracy is only 5 %.
Key Point This confirms
the fact that high degree of precision does not guarantee the accuracy. It is
the accurate calibration that makes the accurate measurement possible.
The
precision is composed of two characteristics:−
•
Conformity and
•
Number of significant figures.
a. Conformity
Consider
a resistor having true value as 2385692 Ω, which is being measured by an
ohmmeter. Now, the meter is consistently measuring the true value of the
resistor. But the reader, can read consistently, a value as 2.4 M Ω due to
nonavailability of proper scale. The value 2.4 M Ω is estimated by the reader
from the available scale. There are no deviations from the observed value. The
error created due to the limitation of the scale reading is a precision error.
The
example illustrates that the conformity is necessary, but not sufficient
condition for precision. Similarly, precision is necessary but not the
sufficient condition for accuracy.
Key Point An accurate
instrument should be precise but a precise instrument may not be accurate.
Significant Figures
The
precision of the measurement is obtained from the number of significant
figures, in which the reading is expressed. The significant figures convey the
actual information about the magnitude and the measurement precision of the
quantity.
For
example, a resistance of 110 Ω, specified by an instrument may be closer to 109
Ω or 111 Ω. Thus there are 3 significant figures. While if it is specified as
110.0 Ω then it may be closer to 110.1 Ω or 109.9 Ω. Thus there are now 4
significant figures.
Key Point Thus more the
significant figures, the greater is the precision of measurement.
Number
of times, the large numbers with zeros before a decimal point are used to
approximate populations or the amounts of money. For example, the price of a
vehicle is reported as ₹ 450,000. This means the true value of the vehicle lies
between ₹ 449,999 and ₹ 450,001. Thus, there are six significant figures. But
what is the meaning of the reported price is, it is closer to ₹ 450,000 rather
than ₹ 440,000 or ₹ 460,000. In this case, there are only two significant
figures. To avoid this confusion, the large numbers are expressed in a
scientific notation using the powers of ten. For example, the price of ₹
450,000 must be expressed as 4.5 × 105. Thus now, there are only two
significant figures. The uncertainty due to the zeros to the left of the
decimal point is usually resolved by such scientific notation.
The
precision can be mathematically expressed as
P
= Precision
Xn
= Value of nth measurement
= Average of the set of measured values
Example
1.5.1 The table shows the set of 5 measurements recorded
in a laboratory. Calculate the precision of the 3rd measurement.
PPPPPPPPPPPP
Solution
:
The
average value for the set of measurements is,
PPPPPPPPPPPP
The
value of 3rd measurement is Xn = 52 where n = 3
PPPPPPPPPPP
This
is the precision of the 3rd measurement.
3. Error
The
most important static characteristics of an instrument is its accuracy, which
is generally expressed in terms of the error called static error.
Key Point The algebraic
difference between the indicated value and the true value of the quantity to be
measured is called an error.
Mathematically
it can be expressed as, .
e = At − Am
where
e = error
Am
= measured value of the quantity
At
= true value of the quantity
In
this expression, the error denoted as e is also called absolute error.
The absolute error does not indicate precisely the accuracy of the
measurements. For example, absolute error of ± 1 V is negligible when the
voltage to be measured is of the order of 1000 V but the same error of ± 1 V
becomes significant when the voltage under measurement is 5 V or so. Hence,
generally instead of specifying absolute error, the relative or percentage
error is specified.
Mathematically,
the relative error can be expressed as,
PPPPPPPPPP
The
percentage relative error is expressed as,
PPPPPPPPPP
From
the relative percentage error, the accuracy can be mathematically
expressed as,
PPPPPPPP
Where
A = relative accuracy
and
a = A × 100 %
Where
a = percentage accuracy
The
error can also be expressed as a percentage of full scale reading as,
Error
as a percentage of full scale reading = PPPPPPP
where
f.s.d. = Full scale deflection.
Example
1.5.2 The expected value of the voltage to be measured is
150 V. However, the measurement gives a value of 149 V. Calculate (i) Absolute
error; (ii) Percentage error; (Hi) Relative accuracy; (iv) Percentage accuracy
and (v) Error expressed as percentage of full scale reading, if the scale range
is 0 − 200 V.
Solution
:
The
expected value means true value, At = 150 V
The
measured value is given as 149 V.
∴ Am = 149 V
i)
e = absolute error = At − Am
= 150 − 149 = + 1 V
ii)
PPPPPPP
iii)
PPPPPPP
iv)
% a = A x 100 = 0.9933 x 100 = 99.33 %
v)
% error expressed as percentage of full scale reading is,
PPPPPPPPPPP
as f.s.d. is 200 V = 0.5 %
4. Sensitivity
The
sensitivity denotes the smallest change in the measured variable to which the
instrument responds.
It
is defined as the ratio of the changes in the output of an instrument to a
change in the value of the quantity to be measured.
Mathematically
it is expressed as,
∴ Sensitiviy = infinitesimal
change in output/infinitesimal change in input
PPPPPP
∴ Sensitiviy = ∆q0/∆qi PPPPP
Thus,
if the calibration curve is linear, as shown in the Fig. 1.5.1 (a), the
sensitivity of the instrument is the slope of the calibration curve.
If
the calibration curve is not linear as shown in the Fig. 1.5.1 (b), then the
sensitivity varies with the input.
PPPPPPPPPPP
The
sensitivity is always expressed by the manufacturers as the ratio of the
magnitude of quantity being measured to the magnitude of the response.
Actually, this definition is the reciprocal of the sensitivity is called inverse
sensitivity or deflection factor. But manufacturers call this
inverse sensitivity as a sensitivity.
Inverse
sensitivity = deflection factor
∴ Deflection factor = 1/ sensitivity
= ∆qi/∆q0 PPPPPPPPPPP
The
units of the sensitivity are millimeter per micro−ampere, millimeter per ohm,
counts per volt, etc. while the units of a deflection factor are micro-ampere
per millimeter, ohm per millimeter, volts per count, etc.
The
sensitivity of the instrument should be as high as possible and to achieve this
the range of an instrument should not greatly exceed the value to be measured.
Example 1.5.3 A particular ammeter
requires a change of 2 A in its coil to produce a change in deflection of the
pointer by 5 mm. Determine its sensitivity and deflection factor.
Solution
:
The input is current while output is deflection.
Sensitivity
= change in output/ change in
input = 5mm/2A = 2.5 mm /A PPPPPPP
Deflection
factor = 1/ sensitivity = 1/2.5 = 0.4 A
/ mm PPPPPP
5. Resolution
It
is the smallest increment of quantity being measured which can be detected with
certainity by an instrument.
Key
Point Thus, the
resolution means the smallest measurable input change.
So
if a non zero input quantity is slowly increased, output reading will not
increase until some minimum change in the input takes place. This minimum
change which causes the change in the output is called resolution. The
resolution of an instrument is also referred to as discrimination of the
instrument. The resolution can affect the accuracy of the measurement. .
Example
1.5.4 A 30 cm scale has 30 uniform divisions. l/20 th of a
scale division can be estimated with a fair degree of certainty. Determine the
resolution of the scale in mm.
Solution : 1 scale division = full scale
deflection/number of division = 30cm/30 = 1cm = 10 mm
Resolution
= 1/20 × scale division = 1/20 × (10 mm) =
0.5 mm
6. Theshold
If
the input quantity is slowly varied from zero onwards, the output does not
change until some minimum value of the input is exceeded. This minimum value of
the input is called threshold.
Key Point
Thus, the resolution is
the smallest measurable input change while the threshold is the smallest
measurable input.
7. Linearity
The
instrument requires the property of linearity that is the output varies
linearly, according to the input. The linearity is defined as the ability to
reproduce the input characteristics symmetrically and linearly. Graphically
such relationship between input and output is represented by a straight line.
The
graph of output against the input is called the calibration curve.
Key
Point The linearity
property indicates the straight line nature of the calibration curve.
The
linearity is defined as the maximum deviation of the actual calibration
curve (output) from the idealized straight line, expressed as a percentage of
full scale reading or a percentage of the actual reading.
The
Fig. 1.5.2 shows the actual calibration curve and idealized straight line.
PPPPPPPPPP
Thus,
the linearity is defined as,
PPPPPPPPPPPP
It
is desirable to have an instrument as linear as possible as the accuracy and
linearity are closely related to each other.
8. Zero Drift
The
drift is the gradual shift of the instrument indication, over an
extended period during which the value of the input variable does not change.
The
zero drift is defined as the deviation in the instrument output with
time, from its zero value, when the variable to be measured is constant. The
whole instrument calibration may gradually shift by the same amount.
There
are many environmental factors which affect the drift. These factors are stray
electric field, stray magnetic field, temperature changes, contamination of
metal, changes in the atomic structure, mechanical vibrations, wear and tear,
corrosion, etc.
The
drift is undesirable and cannot be easily compensated for. It must be carefully
guarded against by continuous inspection.
9. Reproducibility
It
is the degree of closeness with which a given value may be repeatedly measured.
It may be specified in terms of units for a given period of time.
The
perfect reproducibility indicates no drift in the instrument.
The
repeatability is denied as variation of scale reading ans is random in nature.
Both reproducibility and the repeatability are a measure of the closeness with
which a given input may be measured again and again. The Fig shows the input
and output relationship with positive and negative repeatability.
PPPPPPPPPPPPPP
10. Stability
The
ability of an instrument to retain its performance throughout its specified
operating life and the storage life is defined as its stability.
11. Tolerance
The
maximum allowable error in the measurement is specified interms of some value
which is called tolerance. This is closely related to the accuracy.
Actually
tolerance is not the static characteristics of measuring instrument but it is
mentioned because in some instruments the accuracy is specified interms of
tolerance values.
Key
Point The tolerance
indicates the maximum allowable deviation of a manufactured component from a
specified value.
12. Range or Span
The
minimum and maximum values of a quantity for which an instrument is designed to
measure is called its range or span. Sometimes the accuracy is
specified in terms of range or span of an instrument.
13. Bias
The
constant error which exists over the full range of measurement of an instrument
is called bias. Such a bias can be completely eliminated by calibration. The
zero error is an example of bias which can be removed by calibration.
14. Hysteresis
If
the input to the instrument is increased from a negative value, the output also
increases. This is shown by curve 1 in the Fig. 1.5.4. But if the input is now
decreased steadily, the output does not follow the same curve but lags by
certain value. It traces the curve 2 as shown in the Fig. 1.5.4. The difference
between the two curves is called hysteresis. The maximum input
hysteresis and the maximum output hysteresis are shown in the Fig. 1.5.4. These
are generally expressed as the percentage of the full scale reading.
PPPPPPPPPPPP
15. Dead Space
In
some instruments, it is possible that till input increases beyond certain
value, the output does not change. So for certain range of input values there
is no change in output. This range of input is called dead space. This
is shown in the Fig. 1.5.4. There is possibility that instrument without
hysteresis may show the dead space in their output characteristics. Backlash in
gears is a good example which causes the dead space.
16. Span Drift or Sensitivity Drift
If
there exists a proportional change in the indication, all along the upward
scale then the drift from nominal characteristics is called span drift
or sensitivity drift.
The
characteristics with span drift are shown in the Fig. 1.5.5.
PPPPPPPPPPP
Fig.
1.5.5 Span drift
Example 1.5.5 A voltmeter
reads 111.5 V. The error taken from an error curve is 5.3 %. Find the true value
of the voltage.
Solution
: Am = 111.5 V, % e = 5.3 %
PPPPPPPPPPP
∴ 0.053 At = At − 111.5 i.e.
At = 117.74 V
Example
1.5.6
A
0 − 100 V voltmeter has 200 scale divisions which can be read to 1/2 division.
Determine the resolution of the meter in volt.
Solution
:
1
scale division = Full scale division/ Number of divisions = 100/200 = 0.5 V PPPPP
∴ Resolution = 1/2 ×
scale division = 1/2 × 0.5 = 0.25 V PPPPPP
Example
1.5.7
For
the network shown in the Fig. 1.5.6, find the voltage reading on voltmeter, if
voltmeter sensitivity is 1 kΩ/volt. If the voltmeter is replaced by another
voltmeter having sensitivity 25 kΩ/volt, find the new reading. Comment on the
answer.
PPPPPPPPPPP
Solution:
PPPPPPPPPPPP
By
voltage divider, the voltage across the points A − B is,
VAB
= 50 kΩ × 150 /(100 K + 50 K) = 50
V PPPPPP
Now
voltmeter one has sensitivity 1 kΩ/volt, hence resistance offered by the
voltmeter is,
R
= (1 kΩ/volt) × 50 = 50 kΩ
Hence
circuit becomes,
∴ VAB =
(50 kΩ || 50 kΩ) x-PPPPPPPPPPPP
Thus
the percentage error in the reading is,
%
e = true value − measured value/ true value × lOO = (50 − 30)
/50 × lOO = + 40 % PPPPPPPPPP
Now
the voltmeter is replaced by another one having sensitivity 25 kΩ/volt. Thus it
will offer the resistance,
PPPPPPPPPPPPP
R
= 25 kΩ/volt × 50 = 1250 kΩ
Hence
circuit becomes,
∴ VAB =
(50 kΩ || 1250 kΩ)
PPPPPPPPPPP
=
48.077 × 150/(100 + 48.077) = 48.701 V PPPPPP
Thus
the percentage error in the reading is,
%
e = (50 – 48.701)/50 × 100 = + 2.59 % PPPPPP
Key
Point
Thus
the voltmeter with low sensitivity shows more error while the voltmeter with high
sensitivity shows less error.
Example
1.5.8
A
voltmeter having a sensitivity of 1.5 kΩ/volt reads 80 V on its 150 V range,
when connected across an unknown resistor in series with a milliammeter. The
ammeter reads 15 mA. Calculate
i)
Apparent resistance
ii)
Actual resistance of unknown resistor
iii)
Error due to loading effect of voltmeter
iv)
Percentage relative accuracy.
Solution
:
PPPPPPPPP
The
circuit diagram is shown in the Fig. 1.5.7.
The
total circuit resistance, neglecting resistance of milliammeter is,
RT
= V/I = 80/15 × 10−3 = 5.333
kΩ PPPPPPPPP
i)Thus
the appearent value of the resistance is
Rapp
= 5.333 kΩ
ii)
Let us calculate the actual Rx.
The
resistance of the voltmeter be Rv.
∴ Rv = 1.5 kΩ/volt × 150 as 150 V is
full scale reading = 225 kΩ
Thus
PPPPPPPPPPPPP
∴ Rx + 225 = 42.19 Rx i.e.
Rx = 4.462 kΩ
This
is the actual value of the unknown resistance.
iii)
PPPPPPPP
iv)
The relative accuracy.
%
A = (1 − |error|) × 100 = (1 – 0.0236) ×
100 = 97.63 %
Example
1.5.9
A
step input of 5 A is applied to an ammeter. The pointer swings to 5.18 A and
finally comes to rest at 5.02 A.
a)
Determine the overshoot of the reading in amperes in percentage of final
reading.
b)
Determine the percentage of error in the instrument.
AU:
Dec. - 04
Solution:
The
overshoot of 5.18 − 5.02 = 0.16 A when final reading is 5.02 A.
a)
% overshoot = 0.16/5.02 × 100 = 3.187 % ppppppp
b)
% e = (At − Am)/At × 100 = (5 − 5.02)/5 × 100 = −0.4 % PPPPP
Review
Questions
1.
Define any three parameters of the static characteristics of instruments.
AU:
May - 06, Dec -18, Marks 6
2.
Explain the static characteristics of a measurement system, with examples.
AU:
May-11, 13, 15, 16, 18, 19, Dec.-08, 09, 11, 12, 14, 15, Marks 16
3.
A moving coil voltmeter has a uniform scale with 100 divisions, the full scale
reading is 200 V and 1/10 of scale division can be estimated with a fair degree
of certainity. Determine the resolution of the instrument in volt.
[Ans.
: 0.2 V]
4.
A digital voltmeter has a read out range from 0-9999 counts. Determine the
resolution of the instrument in volt when the full scale reading is 9.999 V.
[Ans.
: 1 mV]
5.
A true value of voltage across resister is 50 V. The instrument reads 49 V.
Calculate
i)
Absolute error
ii)
Percentage error
iii)
Percentage accuracy
[Ans.
: 1 V, 2%, 98%]
6.
Define accuracy and reproducibility of an instrument and explain.
AU:
May-14, Marks 6